If a solution contains 70% Ethanol, how much would this equate to in ug/ml or ppm.

Or if a solution contains 5000 ug/ml of ethanol, how much would that be in %?
It's a trick question. You need to specify if the % is w/v or w/w or v/v

That's why concentration should never be given in % or ppm.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
tom jupille wrote:
It's a trick question. You need to specify if the % is w/v or w/w or v/v

That's why concentration should never be given in % or ppm.


V/V
Your second question is easier -- but the answer may not be meaningful.
5000 ug/mL = 5 mg/mL = 0.005g/mL
To get V/V, you need to convert that 0.005 g to mL, so you divide by the density of ethanol (0.8 g/mL) which gives you 0.006 = 0.6% . At that low a level, any volume contraction on mixing will be negligible (see next paragraph).

The first question is more complex because of the volume contraction when you mix ethanol and water. If you ignore that, the calculation is simply the reverse of the one above. Each mL of solution contains 0.7 mL of ethanol. Multiply by the density, and that gives you about 0.6 g per mL. Note that all of these are one-significant-figure calculations. If that's all you need, you're fine; the volume contraction will only be a few percent. If you need more precision, then you have to factor in that volume change on mixing, which means that you need to know whether the "70%" refers to the volumes mixed or to the volume fraction in the final mixture.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
4 posts Page 1 of 1

Who is online

In total there are 6 users online :: 0 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 6 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry