Hi everyone. The autosampler dropped a vial during its run last night and put the system into "abort", all as it should. However this morning I cannot reset the injector, the transport arm is extended and the vial gripper part is in the "hollow" in front of the needle. Initiating the reset function (from chemstation) appears to do something but the arm wants to move to the right and is stopped by the plastic housing round the needle/needle seat area. This results in an "unknown error" message appearing on the chemstation screen.

Powering the ALS off and on doesnt sort the problem.

Does anyone know if I can manually move the arm and gripper "joint", I have tried using a small amount of force but nothing wants to move (either with the power on or off).

Any other suggestions welcome!

Thanks

Chris
You can manually move the gripper arm by reaching in and moving the worm gears at the back of the sampling unit. Move the arm so it has clearance around it before you restart it. Also, pull the tray out before you start doing this.
Time flies like an arrow. Fruit flies like a banana.
Many thanks, that worked a treat and the ALS has reset properly. I had actually tried moving that worm rod but suspect I was twisting it the wrong way. What an idiot :oops: :oops: :oops:

Thanks again

Chris
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry