I've been running this system since 2004 on Chemstation ver D.01... I need SIM/Scan mode in the near future. So I've upgraded to Chemstation E.02.01.1177 and I've been running with it since about mid November 2018.
The problem: About every 4th time I switch to tune and vacuum view to do a quicktune the application hangs when I switch back to instrument control. The only two menu choices are "Abort" and "Exit". Only "Exit" does anything. Then I have to restart Chemstation.
1) Is this a common problem for Chemstation ver E?
2) Could this be because I have to use Bootp to run Chemstation?
I should add that I need to use Bootp because my SmartCard3 has a burnt out data cable connection at the card; so I cannot use the keypad and screen to set the communications for the 5973 inert MS.
Finally, a grump. I really hate how they added icons for everything and especially how they integrated the sequence editor into the application.