Empower Summary Report

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

2 posts Page 1 of 1
Hi ,
Any way of seeing the summary report (% RSD of n=6 for example for precision) before the sample set has been processed
ie all injections complete and now you want to check our sst and standard verification
All this is done using labels and summary report but you have to process sample set first creating a result set
Can you not review peak RSD before processing ?
Hi
There is no any peak data (Rt, area, height) before processing. If you want checking the SST you should load the injections in the Review and process manually. The peak RSD% can be calculated in Excell or any other stat. sw.
Or the Empower can be programmed to process and check the SST samples and stop the sample set before injecting the real samples when the SST fails. You can find the technical note about it on waters.com
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Majestic-12 [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry