Hi ,

How do you determine the points across peak for Peak purity assessment on Empower

Current detector settings
>0.1 min (2 s response time (2.5Hz)

is this 20 ?
Hi ydna1977,

What is/are the width(s) of the peaks of interest at the base--particularly for the smallest/narrowest peak of interest? We'd need to know this as well as the data rate. For 2.5 points/sec, the raw peak should be This calculation is incorrect, please note the EDIT below.

20 points * 2.5 points/sec= 50 sec at the base.

Also, if the installation of Empower has the system suitability module, you'll be able to have these values calculated for you.

EDIT: My apologies--I miscalculated, I'm surprised no one has called me out on it by now.

The calculation should be:

20 points / 2.5 points/sec = 8 sec.

Thus, a peak 8 seconds wide at the base would contain 20 points at a data rate of 2.5 Hz (where Hz = points per sec).
MattM
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry