Empower - amount result issue
Posted: Mon May 28, 2018 7:12 am
by ydna1977
Hi ,
Amount in result set is correct
When signed off the amount is different in component summary tables
Sample table was altered for calibrate , quantitate error but doesn't explain why reporting incorrect amount if no calculations involved
Thanks
Re: Empower - amount result issue
Posted: Tue May 29, 2018 12:54 pm
by ratan234
Hello
I was going through the release notes of Empower3 FR4. There was a PCS # 54145 fixed in that version.
Previously, if you changed the vial amounts associated with a sample set method in the Alter
Sample window after results were generated, the sample set method associated with the altered
results was incorrectly reported in Report Publisher. Now, when you change the vial amounts in the Alter Sample window after results are generated, the associated sample set method contains the correct vial amounts in the report.
Document Reference of release notes: 716004779
Re: Empower - amount result issue
Posted: Thu May 31, 2018 10:27 am
by ydna1977
Hi ,
Bracketing standards used throughout sample set
first 3 brackets are different dose
last brackets are higher dose
Result set amounts are fine
Method report one dose amounts ok , other are lower and not as in result set