Empower - amount result issue

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
Hi ,

Amount in result set is correct
When signed off the amount is different in component summary tables

Sample table was altered for calibrate , quantitate error but doesn't explain why reporting incorrect amount if no calculations involved

Thanks
Hello

I was going through the release notes of Empower3 FR4. There was a PCS # 54145 fixed in that version.

Previously, if you changed the vial amounts associated with a sample set method in the Alter
Sample window after results were generated, the sample set method associated with the altered
results was incorrectly reported in Report Publisher. Now, when you change the vial amounts in the Alter Sample window after results are generated, the associated sample set method contains the correct vial amounts in the report.

Document Reference of release notes: 716004779
Hi ,
Bracketing standards used throughout sample set
first 3 brackets are different dose
last brackets are higher dose
Result set amounts are fine

Method report one dose amounts ok , other are lower and not as in result set
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry