How to change sample type in Shimadzu GCMS Browser software?
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
2 posts
Page 1 of 1
I have ran a calibration followed by a few samples but when the run finished, I realized that I forgot to change the sample type field from "Unknown" to "Standard" in the batch. Does anybody know how to change it after the files have been acquired?
-
- Posts: 2
- Joined: Fri Nov 24, 2017 8:25 am
You can't change the sample type field in GCMS Browser, but you can do the following:
1) Run GCMS Postrun, and go to the Postrun Batch section
2) Open the original batch file that you acquired.
3) Change sample type as required.
4) Click Start to reprocess the batch, using the data already acquired.
1) Run GCMS Postrun, and go to the Postrun Batch section
2) Open the original batch file that you acquired.
3) Change sample type as required.
4) Click Start to reprocess the batch, using the data already acquired.
2 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science