How to change sample type in Shimadzu GCMS Browser software?

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

2 posts Page 1 of 1
I have ran a calibration followed by a few samples but when the run finished, I realized that I forgot to change the sample type field from "Unknown" to "Standard" in the batch. Does anybody know how to change it after the files have been acquired?
You can't change the sample type field in GCMS Browser, but you can do the following:
1) Run GCMS Postrun, and go to the Postrun Batch section
2) Open the original batch file that you acquired.
3) Change sample type as required.
4) Click Start to reprocess the batch, using the data already acquired.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 234 on Tue Feb 06, 2018 7:33 am

Users browsing this forum: No registered users and 1 guest