OpenLAB CDS Chemstation Calibration Table Issue

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
After running a sequence with set calibration levels the resulting automatically-generated calibration table is populated with incorrect peak area values. The update response factor/retention time columns are set to "replace" so it's not averaging anything with past data. If we delete the calibration table and recreate it manually, it populates with correct peak areas. Is there a setting being missed here?

Also, as an aside, is there an intelligent reporter expression that references the dilution factor for a given injection?
I don't know if this applies to your situation, but in chemstation sequences we ALWAYS had the first line (standard) as "replace" with 1 or 2 injections then the second line was same vial of standard but set as "average".

"Replace" cleared any past calibration areas.
In the method settings, you get to choose how and when to update the cal standards and when to use the recently integrated data to do so. Also, you select which "sample" consists of stds and which are actual samples. Any basic errors in these method parameters could produce the results you have observed.
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Andy F and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry