Hi, does anybody have knowledge/experience of using OpenLab CDS to create a custom calculation. I want to compute the %Recovery of a system check sample against the reported/quantitated amount obtained for a calibration standard.
I was thinking this would be quite straightforward however i am failing miserably and would much appreciate any help offered.
Thanks
Re: Agilent OpenLab2 Custom calculations
Posted: Tue Apr 17, 2018 1:21 pm
by ydna1977
Hi ,
Been a while since I have used Open lab
% Recovery should not need a custom field if you add weights and dilutors /multipliers to sample and standards v calib curve (area v amount for stds)
reference worth looking at :
https://www.agilent.com/cs/library/user ... sGuide.pdf
Re: Agilent OpenLab2 Custom calculations
Posted: Wed Apr 25, 2018 8:38 pm
by Camelia
Hi, thanks for the reply. I had previously found the manual you referenced quite useful when working with EZChrom CDS. I am no further ahead in using Openlab2.2 to perform this calculation: Reported amount cal std 2 when run as a sample / The reported cal std 2 amount when run as standard and quantified against the curve * 100.
All ideas are appreciated. Thanks
Re: Agilent OpenLab2 Custom calculations
Posted: Mon May 07, 2018 2:11 pm
by varossf
Hi Camelia, i have a Custom Calculation that does that for you. Of course on demo data but if you send me an email then I could help you.