androidlando wrote:
Hi folks!
I mistakenly ran a calibration curve under sample type "sample" instead of "calibration." Is there a way to remedy this?
I've tried loading the batch and changing the sample type in offline mode but it would yell at me. It says I can only change it to "removed" or the original sample type.
Thanks
You don't detail which version of ChemStation. I have a ton of experience with revision A, none with B, and some with C with OpenLab.
That said: in Revision A, one could simply change to "Calibration" in the Sample Table", and then
"Reprocess Only" using Sample Table Information (there's a box to check to do this). This did work.
I never did anything with Batch Processing.