MS is connected. but not GC. can trigger the injection, but abort and not acquiring any data after injection.

the loss of connection was sudden and not any physical change made to the system.

has a gerstel MAS system linked too.
jiangds06 wrote:
MS is connected. but not GC. can trigger the injection, but abort and not acquiring any data after injection.

the loss of connection was sudden and not any physical change made to the system.

has a gerstel MAS system linked too.


I had a 5975 do this before and it was the remote cable that connects from the back of the MS to the GC, looks like a DB9 Serial port on the MS end. The cable was never moved but somehow it went bad, replacing it fixed the problem.
The past is there to guide us into the future, not to dwell in.
Thank you so much for your replay. but how did you find out? by replacing one by one?
does your chemstation GC portion show as offline in chemstation?

and I can see MS temperatures show up on the chemstation monitor.

James_Ball wrote:
jiangds06 wrote:
MS is connected. but not GC. can trigger the injection, but abort and not acquiring any data after injection.

the loss of connection was sudden and not any physical change made to the system.

has a gerstel MAS system linked too.


I had a 5975 do this before and it was the remote cable that connects from the back of the MS to the GC, looks like a DB9 Serial port on the MS end. The cable was never moved but somehow it went bad, replacing it fixed the problem.
jiangds06 wrote:
does your chemstation GC portion show as offline in chemstation?

and I can see MS temperatures show up on the chemstation monitor.

James_Ball wrote:
jiangds06 wrote:
MS is connected. but not GC. can trigger the injection, but abort and not acquiring any data after injection.

the loss of connection was sudden and not any physical change made to the system.

has a gerstel MAS system linked too.


I had a 5975 do this before and it was the remote cable that connects from the back of the MS to the GC, looks like a DB9 Serial port on the MS end. The cable was never moved but somehow it went bad, replacing it fixed the problem.


I can't remember if the GC would go offline, but all the MS reading were normal, but once it tried to inject the run would either not start or act as if it completed as soon as it injected.

The Agilent tech suggested that as the fix, I thought it was crazy but it turned out that was the problem. I never figured out how as standard serial cable that never moves would go bad, but somehow it did.
The past is there to guide us into the future, not to dwell in.
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry