Hi ,
I have played about with some DAD data on Empower (extracting wavelength etc)
Now every time I review a sample set the Method and processing method for DAD is default and PDA (Bottom right of Empower screen ), even when system is not PDA or GC or GCHS

I ve tried a number of things to correct this with no success
I cannot process either , when review result set it just has = sign and sample set name ,, no data

Thanks
hi think I figures it out

not sure how I did this but here is the solution :

From the Options menu select Method Set Options


The Method Set Options dialogue box opens. This lets you pick a Method Set that will be automatically applied to the data when you open Review. The Method Set you select must contain a derived channel for a specified wavelength (or MaxPlot) and a Processing Method for this to work properly. The first option tells Empower to open and apply the Acquisition Method Set, the Method Set used when you injected the samples.

This was set to DAD as default
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry