Empower 3RF4 impurities calculations

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

4 posts Page 1 of 1
Can you please help me with my problem?!
I don't know how to calculate impurities in my sample if my formula uses the Average Area of 5 consecutive injections of a external Standard, using Empower calculations.
I need % w/w calculation.

To clarfy, I managed to calculate impurities in my sample using %Area, %Adjusted Area, or reffering to the main component but don't know how to use different injection to calculate imps in my probe.

So lost, thank you all for help in advance!
Hi

function for 5 standard should be 'inject standards' and Normal

Clear calibration at start and quantitate at the end of sample set

Add in dilution , sample weights and purity

process sample set should give calibration area (n=5) v amount

Hope this helps
thank you very much for your input. As I understand you explained how assay is calculated, but my question was about calculation of impurities in my sample.

thank you anyways!
Hi ,

once above completed (Including calibrate ) in impurity tab add impurity response (% area , % w/w) and main component to quantify against
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry