Hi,

I'm hoping someone can help me!!

I am running a simple assay as follows:

Std A
Sample X W1
Sample X W2
Std B

I can work out the amount for the samples - and other calculations.
I can also work out the average peak area for Std A and B.
But where my problem lies is with Normalised Area for the standards.
The procedure is looking for the % RSD of normalised area.
I have created a peak based custom field which is simply the peak area of standard A multiplied by the dilution (which is added in the run sample table) and I have called it NormalisedArea_A.
I have then done the same for standard B.
The results of these calculations are fine and correct.

The trouble starts when I try to calculate the % RSD. I cannot do this by component summary as there are two components.

I have tried to combine the custom field for the calculation - in other words apply the custom field for normalised area to both standards and when I do this I only get the correct answer for standard A but not B. I was hoping this would work, so I could use component summary to calculate the % RSD.

I then tried to create a result field based on the % RSD function in empower, however no matter what combination of the peak custom fields for NormalisedArea_A and NormalisedArea_B I use I cannot generated an answer.

It is possible that I am missing a bracket or a dot or something simple, but if anyone can shed some light on this I would be delighted.

Thanks,

Marion :roll:

Instead of creating custom fields for "NormalisedArea_A" and "NormalisedArea_B" you might try creating a field called "NormalisedArea" and just multiply the peak area by the dilution factor for anything. Then, in your report you can just report results for A and B standards.

If you paste the script then I could look at that too.

Thanks for your help. Here's what I've done:

Peak Area Std A = 11880920 Dilution = 0.9970 (Normalised Area = 11845277)

Peak Area Std B = 11839171 Dilution = 0.99980 (Normalised Area = 11836803)

The fields I have created are:

Normalised Area_A = (S0101.1.(Area))*Dilution - I get the correct answer here.

Normalised Area_B = (S020B.1.(Area))*Dilution - I also get the correct answer here and can report them separately. But can't do %RSD

This morning I tried (S0.1.(Area))*Dilution and filtered the table to only report injections with labels S0* - I got no answers!

Then I tried Area*Dilution and still got no answers!

I'm not sure what I'm doing wrong. I understand that it would be better to calculate the value for all and then only report what I need. I'm not sure where I'm going wrong!

OK, I restated Empower and reprocessed my sample set and I have generated the correct values. :lol:

Thank you so much for your help, I am currently working through all the QC analysis that use excel....so I will probably be back!
4 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry