I can't change the loop size in chromeleon
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
5 posts
Page 1 of 1
Hello, can someone helping me please? I can't change the loop size in chromeleon command (it's hided). I have a dionex ICS-3000 with a 25ul sample loop and want to change with a 10ul but i can't enter the new value in control pannel. How can I proceed please?
-
- Posts: 835
- Joined: Thu Apr 14, 2005 7:00 am
i don't have an IC system in my lab, but if it is possible to change the loop size of the sampler, then the new setting has to be made in the server configuration and not in the panel.
-
- Posts: 218
- Joined: Thu Mar 31, 2005 5:33 pm
- Location: USA
Of course you can. The loop size/injection volume goes with your project/method, not with your control panel. You can change it through method edition or the VB method window.
-
- Posts: 193
- Joined: Thu May 11, 2006 11:10 pm
- Location: California, USA
I partly agree with unmgvar, in that the change must be done in the Server. However, you can change the syringe (size) but not the sample loop (size).
I do not have the same system listed by Lyser. However, in CM, I can change the syringe size this way:
Choose Server configuration.
-> Click on "Autosampler"
-> ->Select Tab "Syringe and Stroke Sync"
-> -> -> Click on the Select button "250uL" or any others
-> -> -> ->Click OK.
Hope that this helps.
Alfred
I do not have the same system listed by Lyser. However, in CM, I can change the syringe size this way:
Choose Server configuration.
-> Click on "Autosampler"
-> ->Select Tab "Syringe and Stroke Sync"
-> -> -> Click on the Select button "250uL" or any others
-> -> -> ->Click OK.
Hope that this helps.
Alfred
-
- Posts: 310
- Joined: Mon Aug 30, 2004 4:56 pm
- Location: Sunnyvale California USA
Lyser,
First off, if you changed the sample loop and want your instrumentation setup to reflect this, changing a loop size can be done through the front panel of your AS. Press the menu button until you get to the main menu screen, then press 5 then press 3 which should be used to the plumbing options setup screen where you can enter the loop volume. If you want to change the injection volume in your sequence, it's one of the standard columns in the sequence. You use this to define the volume of sample loaded during the loop loading process.
Chris
First off, if you changed the sample loop and want your instrumentation setup to reflect this, changing a loop size can be done through the front panel of your AS. Press the menu button until you get to the main menu screen, then press 5 then press 3 which should be used to the plumbing options setup screen where you can enter the loop volume. If you want to change the injection volume in your sequence, it's one of the standard columns in the sequence. You use this to define the volume of sample loaded during the loop loading process.
Chris
5 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science