Empower 2 Instrument Failure

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

5 posts Page 1 of 1
Hello,
Maybe someone could help me on that issue on Empower 2 :

Instrument failure (detector is stopped) during a sequence.
In the message center is written : Instrument Failure W2690/5@ 7 (HPLC detector Address)

Do you know the origin of this issue ? is it a hardware or software error ?

thank you

You have problem with your hardware (detector)

What kind of detector are you using?

It's a Waters HPLC 2695 with a 2996 detector.

Do you know how to solve this hardware issue ?

In your quickset, you can run PDA diagnostics to see where the problem is located in the PDA.

Maybe you should run these tests and come back with the results?

It could be a lamp which has excessive usage.


Bart

Or, you have a cleanout method or your method is programmed to shut the detector lamp off well after the end of data collection. Either way, we commonly see this message w/ our Empower installation. It's no big deal, as long as it's expected.
Thanks,
DR
Image
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry