Hello, I was wondering if anyone knows how you can adjust the slope sensitivity in a method on Chemstation, so it does not recognize the background noise as peaks. I am also trying to get the program to recognize the solvent peak to adjust the baseline for peaks that occur above the baseling and be able to save these manual integrations to the method. We do not have the manual or a detailed tutorial to figure this out and so it is difficult to to do this by trial and error. Also if anyone knows where to find a manual or a tutorial for an older version of Chemstation (A.06.03) we tried calling Agilent but since this version is obsolete they are unable to help us.

Thanks for any help,

Andrew Gidora

There are two ways of adjusting the slope sensitivity. firstly, to change one trace, load up the desired trace and then select integrationa nd integgration parameters from the drop down menu. sensitivity should be the top option on the table. the lower the value the more senistive the system will be.once the paramters have been changed of you select integrate from the drop down menu then you can see what effect the changes have. the second option is permantly change the method. this is done by edit entire method from the method drop down menu and go through the various options untill you come across the parameter table (this is the same table that you would get from the previous method). once again adjust the sesnitivity as desired.

Personally i would use the first option to determine what parameters are required and then go into the method and change accordingly.

hope this helps

steve

Thank you this advice helped us, we are able to get rid of he "junk" now and find the peaks we want

Regards,

Andrew
3 posts Page 1 of 1

Who is online

In total there are 5 users online :: 0 registered, 0 hidden and 5 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 5 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry