Empower 3 - two Identical results

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

4 posts Page 1 of 1
Hi ,

Why do I have 2 sets of data when I processed my sample set in Empower ?

I have clear calibrate , calibrate and quantitate and also samples as 'don't process and report ' and standards as normal

Thanks
Did you use a standard label more than once? Example for the first bracket you could have used S0101, did you repeat this label anywhere else in Sample Set? If so you will get duplicate results.
normal standards followed by a calibrate line at the end will result in duplicate standard results
Thanks,
DR
Image
Can you show us a screenshot of the sequence?
4 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry