Empower 3 Impurity Tab
There is an option to select your impurity response in Empower 3, what exactly is the difference between "%w/w", "%Area" and "%Area against Main Component"?
My understanding is:
%Area = area of peak as a function of the Total area of all peaks in the chromatogram, both system peaks and sample peaks eg peak of impurity = 2000. Total area of every single peak in the chromatogram = 214000. % Area = (2000/214000)*100
%Area against Main component = corrected area of peak (area x RRF)/area of active*100
%w/w = I'm not entirely sure!
Can anyone clarify? Thanks.
did u ever figure this out? because i am only starting to use Empower 3 for impurities calculations, and have the same (and more) questions?!
% w/w calculated from calibration curve of standards , dilution factor and weights entered into sample set wrt main component
think back to first principles ie
sample area/std area * std weight/dil* dil f sample /sample weigh *purity
Most users ever online was 234 on Tue Feb 06, 2018 7:33 am
