Chemstation version: B.03.01 SR1

Is it possible to stop the Quat pump on a Agilent 1100 via a Chemstation Macro? If so can this macro be initiated via a command from another program? Eg: Excel

Thanks in advance
Under Sequence Parameters, check box, then choose PumpAll Off? That's what we do.
Not quite what i was looking for but thanks anyway.

I want to be run / send a command to Chemstation to shut down the pump, is this possible from a batch file?
http://www.chem.agilent.com/Library/use ... anding.pdf
page 178, refers to a macro, plus other options.
e.g if you know your system is going to run for 7 hrs, make a method that runs for 7 hrs,
then make a sequence with the 7hr method and place the shutdown function to occur as suggested by KM-USA.

External control can be done by using DDE (Dynamic Data exchange), the DDE handle changes
whenever Chemstation is shut down and restarted, which is a real pain.

Alex
If you want to turn off the pump at a set time, select Start > Agilent ChemStation > Scheduler. In Scheduler, enter the date and time, and the command is: "PumpAll Off"

You can select 'Do Once', 'Do Daily',.. Weekdays, .. Weekly.
5 posts Page 1 of 1

Who is online

In total there are 18 users online :: 2 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry