viewing %RSD results empower 3

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

5 posts Page 1 of 1
hi,

is there a way of viewing your %rsd results for a run without having to go into a report?
results tab
Thanks,
DR
Image
fantastic, i wonder is it a field that you have to set up yourself because im not seeing any columns related to %rsd under the results tab?
Assuming you are looking at injections or unprocessed results, you will have to do some online "processing" and add the %RSD column to your table beneath the chromatogram.
Thanks,
DR
Image
I dont think there is a "%RSD" column in the chromatogram view in Empower. You either add it to a column in a report table or you can create a custom field for Empower to calculate the %RSD and this value will populate when you generate a result set. For example, if you wanted the %RSD of amount in all your samples, you could create a formula like U%.%..%RSD(Amount) and add Summarize Custom Fields as the last line of the sample set, and it will then calculate the %RSD of amounts for all samples beginning with the label U...when you are looking at the result set.
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry