Calibration Report Issue Empower 3
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
6 posts
Page 1 of 1
My company is trying to make a single report method for every test that we perform. When trying to report a calibration curve in Empower 3 I choose the result set and open publisher with the established report method the system wants to generate a calibration graph and table for every injection from the run. I've tried only selecting the results that pertain to the cal curve and I still run into the same problem. I've played with every setting I could think of and it doesn't seem to resolve the issue. Is there something I'm missing?
-
- Posts: 2071
- Joined: Tue Aug 17, 2004 7:59 pm
- Location: 38.590967, -90.213236
Method properties: Summary by All
Thanks,
DR
DR
-
- Posts: 3
- Joined: Mon Aug 15, 2022 8:50 pm
I've tried Summery by All and it still gives the same duplicates just in a single report.
-
- Posts: 2071
- Joined: Tue Aug 17, 2004 7:59 pm
- Location: 38.590967, -90.213236
https://blog.waters.com/empower-267-calculating-amount-concentration
Try reviewing this series of tips on calculations...
Try reviewing this series of tips on calculations...
Thanks,
DR
DR
-
- Posts: 301
- Joined: Tue Jul 12, 2016 4:47 pm
Can you not use a data filtering condition in your report table and only use one sample? For example, say you run 10 samples 1A to 10A, filter by SampleName = 10A, that way only one cal curve graph and table is shown but its still relevant for all the samples, presuming all samples are quantitated off this curve.
-
- Posts: 3
- Joined: Mon Aug 15, 2022 8:50 pm
EmpowersBane wrote:
Can you not use a data filtering condition in your report table and only use one sample? For example, say you run 10 samples 1A to 10A, filter by SampleName = 10A, that way only one cal curve graph and table is shown but its still relevant for all the samples, presuming all samples are quantitated off this curve.
That seems to have worked. I'll try it out with different result sets to verify. Thank you!
6 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science