Sample weight not used in concentration calculation

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
Hello!

I am setting up an Agilent 1100 HPLC with ChemStation in a new lab. The LC already has a calibration curve set up, however I am running into an issue during sample data analysis. When I go to view the concentrations of the samples I have run, there is a line on the PDF that says the sample weight was not used in the concentration calculation.

What steps should I take to fix this issue? I'm more than happy to provide any other information if necessary! Thank you in advance for your help
I remember having the same issue back when we got our first 1050 units computerized, with Chemstation A3. All the later 1050 and 1100 systems with A version software (and the GC systems as well) were almost identical for the user.

However, that was about the mid-1990s, and I don't remember all of that. We always used a Sequence Summary Report to calculate the results, and there was something in there that had a box that needed to be checked to use the sample weights in the Sequence Table.
Thank you for your response!

I wasn't able to find a box to check in the Sequence Summary Report - I'm running ChemStation version B.04.03-SP1 [87] if that helps!

These are my current report settings:

Image

Is there anything I should change/anything else I should try?
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry