Whats is the calculation for % Assay?
Posted: Tue Sep 15, 2020 8:50 am
This is causing some disagreement in my workplace, can anyone provide some clarity. So we have a test where we weigh out 100mg of sample and dissolve in 100mls of sample solvent, yielding a 1mg/mL or 1000mcg/mL solution. Standards have the same concentration and same sample solvent. So our test method states that the content of active is basically the same calculation that Empower 3 uses for amount:
Area of sample/mean area of standard * conc of standard (mcg/mL).
This just gives the actual amount in the sample. To get the % Assay, I divide this amount by the theoretical amount and * 100 to get % Assay. Example:
Weigh 100.098mg which yields a conc of 1000.98 mcg/mL. I quantify this sample and get an amount of 1000.14 Therefore 1000.14/1000.98*100 = 99.92% recovery or assay.
Any thoughts on this method?