Chemstation B.04.03 not recording both FID channels, HELP..

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

2 posts Page 1 of 1
Just switched over to B.04.03 version, running Headspace with Dual FID on 5890 Series II.

I use front injector
I have a Y at the inlet and have two columns, one going to FID A and the other to FID B
On the method I select SIG A and SIG B, front injector

In pre run, I can monitor both Detector signals, but once the run starts, for some reason, the computer output is the same (Det A) for both Sig A and Sig B. I can not see the other signal.

I am wondering if I am missing something. If any of you have the same system please advise.

Thanks in advance for any help.
In the software, if you click on Instrument, Edit Parameters then the Signals option. Make sure the "Save Data" box is checked for both FID A and FID B.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry