Hello all!

I'm not sure if I'm just blanking, but I've been trying to get this custom field to work without success. I have a molecule that has two criteria to meet:

1. Height/Area Ratio must be GTE 85% of the Maximum Height/Area Ratio
2. Peak Area must be GTE one third of the maximum peak area

Is there a way for me to write a custom field that can find the maximum within a result set? Or is this only available to do in an injection report?

Here's an example of a custom field I have prepared for the first criterion:

U%..MAX(Peak1[Height_Area_Ratio])*0.85

U is the label for all samples. Height_Area_Ratio is just Height/Area. Peak1 is a default peak name I'm using for this.

I've been trying to see if there is a way to avoid using Excel and just having Empower calculate the maximums of a result set.
I'd look into using a filter instead of a CF for this. It can be a summary row or a separate table, but I'm pretty sure Maximum is a filtering option...
Thanks,
DR
Image
3 CFs will do this. I just tested it on some example data in my lab and it worked fine. Copy formulas exactly as below, you are missing a period (.) in your post

Create a Result, Real, CF with search order Result Set Only and formula:

U%.%..MAX(Peak Name[Height_Area_Ratio])*0.85. This will calculate 85% of the maximum Height/Area ratio in all injections labelled with U...

Call this Result_CF1 (or whatever you prefer)

Then a second Result, Real, CF again search order Result Set Only, formula:

U%.%..MAX(Peak Name[Area])*0.3333 (this will calculate a third of the maximum peak area in all injections labelled with U...

Call this Result_CF2

Create a third Peak, Bool CF, search order Result Set First or Result Set Only, Use as Position, formula:

GTE(Peak Name[Height_Area_Ratio],Result_CF1)*GTE(Peak Name[Area],Result_CF2)

This will test each peak against the 2 conditions you want and return whatever text you put in position 1 of the translation table (Yes) and if both conditions aren't satisfied, this CF will return whatever text you entered for position 0 of the translation table (No)

Because of the search orders, you need to process a sample set to generate a result set for all the CF calculations to kick in they will be visible in the result set and a report method can be created to add in the CFs.
3 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry