Hello:

We added an FID to our 5975/7890 system in order to be able to analyze for TPH using the chemstation software. We are able to get a very good calibration but we are getting very high counts for our blank samples (baseline becomes negative)! Any suggestions.

Thank you very much for your help.
Get your program set up. Fire up the system and press run a few times to clear out the column. Now do a column compensation run. That should get you a flat baseline for your calibration. When you do a run on a new day. Try doing a couple blank runs (by pressing run on the front panel). If it looks good, run an actual blank before the CCV. If the blank is good, proceed. Otherwise back up and do another column compensation run.
2 posts Page 1 of 1

Who is online

In total there are 11 users online :: 0 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry