Hi ,
Anybody prepared or already switched to windows 10 :
•Do I have to upgrade?
•What are the benefits of upgrading now
•Is there an easy process I can follow
What are you switching?
Windows 7 support ends in January 2020 so update to windows 10
Anybody thinking of this ?
All office PCs with connection to the internet must be upgraded for security reasons, of course. Acquisition workstations can only be upgraded if the instrument control software will work under Windows 10. Ask your software supplier for Windows 10 support. If it is not available, separate your workstation from the internet and continue your work as usual.
In my eyes there is no need to upgrade a functional acquisition workstation just to be able to use internet.
bunnahabhain wrote:
All office PCs with connection to the internet must be upgraded for security reasons, of course. Acquisition workstations can only be upgraded if the instrument control software will work under Windows 10. Ask your software supplier for Windows 10 support. If it is not available, separate your workstation from the internet and continue your work as usual.
In my eyes there is no need to upgrade a functional acquisition workstation just to be able to use internet.


I agree, except my instrument acquisition workstation is also my "office" PC where I get the company email which is now "in the cloud" so it requires internet connection. Things would be so much easier if IT could go to the trouble of keeping email in house instead of migrating to the cloud.
The past is there to guide us into the future, not to dwell in.
James_Ball wrote:
...my instrument acquisition workstation is also my "office" PC where I get the company email which is now "in the cloud"...


Does your IT department like malware and viruses?
I ask because that's a great way to get malware and viruses.

Most acquisition PCs require lots of open ports and most data systems do not work well in the presence of anti-malware programs. This results in port vulnerabilities and no effective means of scanning several folders on a given hard drive (or you close the ports and scan the CDS folders and get no data from your instruments).

You should really get another PC for your email and whatnot while leaving the instrument PCs behind a very restrictive firewall/proxy server or just make them part of a subnet with no direct outside access.
Thanks,
DR
Image
DR wrote:
James_Ball wrote:
...my instrument acquisition workstation is also my "office" PC where I get the company email which is now "in the cloud"...


Does your IT department like malware and viruses?
I ask because that's a great way to get malware and viruses.

Most acquisition PCs require lots of open ports and most data systems do not work well in the presence of anti-malware programs. This results in port vulnerabilities and no effective means of scanning several folders on a given hard drive (or you close the ports and scan the CDS folders and get no data from your instruments).

You should really get another PC for your email and whatnot while leaving the instrument PCs behind a very restrictive firewall/proxy server or just make them part of a subnet with no direct outside access.


Most of the acquisition PCs run dual NICs one for the instrument and one for the house network. The internet connections are filtered through the main server as the firewall/proxy which is scanning and recording all inbound and outbound traffic. Certain individuals do not have access to internet when they log into a PC just in case.

In the past it was like pulling teeth just to get a PC that was a little faster to run the instrument, and we usually replaced the instrument PCs with the hand-me-downs from the front office(which makes no sense since a GC/MS or LC/MS/MS will use much more processing power than spreadsheets and email).

But things are changing around here and we now have someone who realizes the importance of the instruments and their needs, but small facilities will often have to make due with what is on hand.

Luckily I am finding that most of the Agilent software will run on W10 even though it isn't certified/supported on it. Though a few legacy instruments need tricks to get them to accept an IP address when they just do not communicate with the BootP as a service.
The past is there to guide us into the future, not to dwell in.
I generally stick with "offline only" for the instrument computers under my control.

I have ONE GC computer where I can connect to WiFi on-demand, but I never leave the connection open. That one is running Windows 10, and it's controlling an SRI GC. Unlike other companies, SRI actually makes their acquisition software freely available on their website, and have a fully validated and ready to go version for Windows 10.

Not too long ago, I "fixed" a 7820/5975 that's not under my direct supervision(it belongs to a research group) but that I get asked to work on as needed(i.e. they break something on it). On this particular occasion, the GC and MS weren't communicating with the computer, and as soon as I looked at it, I knew why-they had attempted to connect it to the internet by running a wire from the ethernet hub that controls the instrument to a wall jack, and although the internet worked the school network had messed with the IP addresses on everything. After I disconnected it and set everything back to default, it was fine.

Interestingly enough, though, I was wearing a different hat today servicing an NMR. One of ours was being stubborn about tuning, and our NMR guy is out of town this week. Those systems all run RHEL, and are internet connected(I think they have dual NICs-one for instrument interface and one for the computer). I actually came to appreciate that as I was reading emailed instructions from the NMR manager and acting on them from the same computer rather than pulling up the email on my phone. I'm not sure whether or not RHEL does Flash(or if these computers have it installed) so about all the internet access on them gets used for is to email results, and probably more often to access the online scheduling system. Since vNMRj is now open source, presumably it won't be a security risk to keep updating both the OS and the software.

The instruments under my supervision are a nice anthology of older versions of Windows. Most of my stuff is on Windows 7, aside from the SRIs on Windows 10, but I have a decent number on XP still, a few on 2000(including the 5890 FID/NPD and a Finnigan LC-MS that's a bit of a work in progress), the 5890/5971 is on 98, and until a couple of things gave up the ghost at the same time I even had a Nicolet FTIR/Raman on Windows 3.11(I'm trying to get the money to both fix the instrument and buy a newer version of the software that will run on Windows 2000). I also can't forget the home-built NMR that runs on Mac OS 9-I provide tech support on that computer since I have relatively "fresh" OS 9 knowledge and also can support the hardware. Offline definitely rules the roost, though.
James_Ball wrote:
Most of the acquisition PCs run dual NICs one for the instrument and one for the house network. The internet connections are filtered through the main server as the firewall/proxy which is scanning and recording all inbound and outbound traffic. Certain individuals do not have access to internet when they log into a PC just in case.

In the past it was like pulling teeth just to get a PC that was a little faster to run the instrument, and we usually replaced the instrument PCs with the hand-me-downs from the front office(which makes no sense since a GC/MS or LC/MS/MS will use much more processing power than spreadsheets and email).

But things are changing around here and we now have someone who realizes the importance of the instruments and their needs, but small facilities will often have to make due with what is on hand.

Luckily I am finding that most of the Agilent software will run on W10 even though it isn't certified/supported on it. Though a few legacy instruments need tricks to get them to accept an IP address when they just do not communicate with the BootP as a service.


While it is a good idea to limit access on instrument PCs, there are still several open ports that are required in order for the acquisition PC to communicate with the server (assuming a C/S environment). This is where the risk comes in. Also, all you need is a badly hung Outlook process to force a reboot and take out several active runs to get things going again. As cheap as PCs are these days, they really should keep a "hands off" policy on instrument PCs.

@ benhutcherson - 3.11 and 2000 machines? You have my condolences. I rather liked 3.11 but was a much bigger fan of 98, XP and 7. I get that 10 works, but I think it is change for change's sake and they always manage to take away things that I used whenever there's a big upgrade. Credit for adding in a virtual drive mounting utility for people who d/l an occasional ISO, but I miss things like photomanager and integrated LAME.
Thanks,
DR
Image
DR wrote:
James_Ball wrote:
Most of the acquisition PCs run dual NICs one for the instrument and one for the house network. The internet connections are filtered through the main server as the firewall/proxy which is scanning and recording all inbound and outbound traffic. Certain individuals do not have access to internet when they log into a PC just in case.

In the past it was like pulling teeth just to get a PC that was a little faster to run the instrument, and we usually replaced the instrument PCs with the hand-me-downs from the front office(which makes no sense since a GC/MS or LC/MS/MS will use much more processing power than spreadsheets and email).

But things are changing around here and we now have someone who realizes the importance of the instruments and their needs, but small facilities will often have to make due with what is on hand.

Luckily I am finding that most of the Agilent software will run on W10 even though it isn't certified/supported on it. Though a few legacy instruments need tricks to get them to accept an IP address when they just do not communicate with the BootP as a service.


While it is a good idea to limit access on instrument PCs, there are still several open ports that are required in order for the acquisition PC to communicate with the server (assuming a C/S environment). This is where the risk comes in. Also, all you need is a badly hung Outlook process to force a reboot and take out several active runs to get things going again. As cheap as PCs are these days, they really should keep a "hands off" policy on instrument PCs.

@ benhutcherson - 3.11 and 2000 machines? You have my condolences. I rather liked 3.11 but was a much bigger fan of 98, XP and 7. I get that 10 works, but I think it is change for change's sake and they always manage to take away things that I used whenever there's a big upgrade. Credit for adding in a virtual drive mounting utility for people who d/l an occasional ISO, but I miss things like photomanager and integrated LAME.


I have to say my best running instrument control PC ever was MSDChemstation running on NT4. I had a few with battery backups that ran over a year between reboots with no slow down or problems at all. Of course there was no USB support, but those were just rock steady machines. W7 running MSDChemstation experiences a memory leak that wasn't there in XP, as about once a week I get the error that "this computer is running out of memory" every time a new run in a sequence starts. Have to reboot to get rid of the error. We are getting ready to upgrade one of our last XP systems to W10, so wish us luck that everything will work, if the thing wasn't causing files to lose ownership and lock us out of them I would not even be thinking of upgrading, but it seems to be a problem with the OS and I really don't have a disk handy to reload it.
The past is there to guide us into the future, not to dwell in.
10 posts Page 1 of 1

Who is online

In total there are 12 users online :: 1 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: Baidu [Spider] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry