I am trying to replace a null value for Amount with Not Detected. I did it in 2 steps:
REPLACE(amount,-9999)
ENUM( ........., EQ(amount,-9999)
In the translation table, "Not Detected" is the value of the enumerated custom field. However, when the peak is missing, amount does not get the value -9999 and the enumerated custom field does not equal, "Not Detected". I thought replace will assign a new value when the field equals Null.
What am I doing wrong?
Thanks,
AA
There is a built in database code for "Not Detected" in Empower. That code is -60002.

If I had to code this, I would personally use:

EQ(Peak Type,"Missing")*-1*60002

In this field, if a named peak is not found, then the result is 1*-1*60002, or -60002 which the database will automatically translate as Not Detected. Use the appropriate REPLACE function to get that Not Detected into the cell you are after.
Thanks. I will try that.
3 posts Page 1 of 1

Who is online

In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry