I am running into a bit of a snag with CE. A lot of my samples are very high in salt mainly NaCl. I work with dried flavor powders. They average 25% weight NaCl but can go even higher. Anyways the high salt is causing terrible peak shape from destacking and significant migration time shifts.

I am running a method for Inosinate, Guanylate, and Thiamine on a 50um-50cm cap using 60 mmol/L borate/2-amino-2-methyl-1-propanol-80 mmol/L sodium dodecyl sulfate (pH 9.6) BGE. My LOQ is 10ppm injection so with a 1/10 dilution my sample LOQ is 100ppm.

Is there anything I can do other than dilute the sample more and lose more sensitivity? I am playing with increasing the borate/AMP to as high as 500mM though that will slow the EOF and may cause heating issues. I am also unsure of the solubility of the SDS in that high salt.
Maybe you could do a dialysis?
I've tried a variety of things so far. I tried the acetonitrile method using 66% acetonitrile in the sample. It reduced the migration time significantly and thiamine disappeared. I suspect it may have crashed out of solution.

I've been trying water plugs, electrokinetic injection etc.

I also tried upping the buffer concentration but it caused issues with the baseline, and of course high current and high migration times due to reduction in the EOF.

I am reading stuff on stacking.

I guess I could worse case scenario do ion exchange SPE but I'd have to do two because thiamine is cationic and Inosinate and guanylate are anionic.
Haven't really found a solution to high salt other than dilute and use the highest ionic strength buffer possible (can't go much higher than 100mM for my method).

For sample cleanup I have had good results with methanol/water:chloroform protein precipitation followed by blow down and redissolution in water. It has the added benefit of removing the fat and many nonpolar food/flavor compounds as well.
4 posts Page 1 of 1

Who is online

In total there are 7 users online :: 0 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry