IC column Dry Out
Posted: Tue Oct 22, 2019 7:34 pm
Hello all,
We were running on a CarboPac 9mm x 250mm column. Due to some instrumental errors, it kept pumping the solvents overnight at 70% water and 30% 1M NaOH. When we discovered it, water has already run out but NaOH hasn't. We primed the pump and flushed the column for 12 hours. Still, we observed peak broadening and tailing from standard injections.
We purge all the eluent bottles with nitrogen, so is the problem caused by the presence of nitrogen gas in the column? Can it be fixed or is the resin permanently damaged?
Thank you for the help!
We were running on a CarboPac 9mm x 250mm column. Due to some instrumental errors, it kept pumping the solvents overnight at 70% water and 30% 1M NaOH. When we discovered it, water has already run out but NaOH hasn't. We primed the pump and flushed the column for 12 hours. Still, we observed peak broadening and tailing from standard injections.
We purge all the eluent bottles with nitrogen, so is the problem caused by the presence of nitrogen gas in the column? Can it be fixed or is the resin permanently damaged?
Thank you for the help!