Have you got any experience with determination SULFATE and SULFITE in sample from technologycal waters. I can use for this app. METROHM vario 940 with SeS suppresion. Can I use Metrosep A Supp 10 250/4 for this app? Have you got experience with standard stability ? eg Will not it decompose SO3 to SO4 during calibration ?


Thank you
BR

Sam
Hello,

In order to separate and to make sure there is no decomposition you need to add about 1-2.5% acetone to your eluent. Also ASUPP 7 250 is the recommended column for this analysis.

-Doug Castrillo
dcastrillo@metrohmusa.com
Metrosep A Supp 10 also works.
The best stabilization of sulfite is adding 10 mmol/L formaldehyde to the sample /Diluent. Also 2% isopropanol showed to work.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Thank you,

I found IC-standard from Fluka (Sigma - Aldrich), 1000 μg/mL in 1% NaOH, do you have any experience with standard ?

Thank you
This standard is stable as it is. For dilution I would recommend to use one of the above mentioned stabilization methods. These dilutions should be renewed quite frequently.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
5 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry