- Posts: 25
- Joined: Fri Sep 07, 2012 8:45 pm
Fluoride matrix effects
Discussions about IC and related topics
4 posts
Page 1 of 1
We have a Metrohm 882 running epa 300.0 and are having problems with quantifying to the expected concentrations of fluoride in certain matrices. Our QC samples (ICV, CCV, MS) consistently pass (+/-10%), but for some samples, we get ~ 50% expected recovery on certain wastewaters, and ~200% expected recovery on certain soils. Any ideas on what could be going on? We cleaned the suppressor with H2SO4/Oxalic acid but that didn't seem to help. MS/MSD of the problem samples get close to 100% recovery for fluoride which is what is confounding.
-
- Posts: 362
- Joined: Tue Aug 31, 2004 6:54 am
Dear Mritari
Samples containing alumina or titanium (as examples) may show low fluoride concentration by IC due to absorption. Only free fluoride is detected.
A too high response is more difficult to explain. Short chain organic acids might interfere.
Do not hesitate to send me chromatograms (*.idet) with details of the procedure (markus.laeubli@metrohm.com).
Samples containing alumina or titanium (as examples) may show low fluoride concentration by IC due to absorption. Only free fluoride is detected.
A too high response is more difficult to explain. Short chain organic acids might interfere.
Do not hesitate to send me chromatograms (*.idet) with details of the procedure (markus.laeubli@metrohm.com).
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
-
- Posts: 3475
- Joined: Mon Jan 07, 2013 8:54 pm
- Location: Western Kentucky
I think Acetate can elute near Fluoride, depending on the column and eluent being used.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 94
- Joined: Wed May 22, 2019 5:19 pm
Acetate is very close and will co-elute. If the F level is high enough the acetate peak will add to the F peak and may not show up as even a bump in the peak. That might explain a high recovery, but not a low one. Running a high dilution of samples allows you to see if there is any co-eluting acetate. My F recovery on high level F decreases as my column ages. It only calibrates well up to 4 ppm when it's old. When new it works well up to 8 ppm. I would try dilutions.
4 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest