Hi all,

I have a Dionex ICS 2100, which has a Rheodyne 9900-013 valve which needs maintenance.

It was quite complicated to find the part number of the rebuil kit, that is 075973

Now I'm trying to get the Stator, I contacted Thermo and IDEX (valve manufacturer) and they passed me the following part number for the stator: 063180 and 044970, however those numbers are not listed on the thermo website and I cant buy it.

A colleague in the US also does not get the stator with those part numbers, could someone tell me what the Stator part number is?

Thank you very much and greetings.
Hi Emefede,

063180 is the correct Thermo p/n for your ICS-2100 injection valve 6-port stator. Please contact your nearest Thermo office or sales rep to purchase this.

The other part you mention (044970) is the stator for a Rheodyne 9126 valve.
Best regards,

John Guajardo
Product Manager, IC Columns
Thermo Fisher Scientific
Hi John,

Thank you so much!!!!

Regards!
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 0 registered, 0 hidden and 2 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 2 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry