Hello,

We are collecting peaks from IC. We have a flow splitter connected to the outlet of column, PAD detector and the fraction collector. We tried to measure the delay time before by placing an empty tubing between the flow splitter and fraction collector, then turning on the pump and measuring the time for liquid to drip out of the fraction collector. But our pump needs around 30s to achieve a stable pressure and the targeted flow rate, which may lead to errors in the measured delay time. Are there any ways to get the delay time in this case?

Thanks
place a valve after the flowsplitter. start the pump in the position without the empty capilary when the pressure is stable, swich to the empty capillary and the measure the time.
You also may calculate the volume of the capillary, then calculate the time with volume and flowrate. The result should be quite good.
Dr. Markus Laeubli
Manager Marketing Support IC
Metrohm AG
9101 Herisau
Switzerland
2 posts Page 1 of 1

Who is online

In total there are 3 users online :: 0 registered, 0 hidden and 3 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 3 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry