Hello,
I ran a seven anion standard mix and the baseline of my chromatogram is not all even. For each of the 7 individual peaks in the standard, each peak would start to rise at 0 uS but as each peak goes back down, the peaks will drop a bit lower than 0 uS and then would slightly go back up again to 0 right before the next peak would come out. Therefore, between each peak there would be an uneven baseline. Also, when I ran this standard about a month ago, I was getting a stronger signal for each of the anion peaks. Would anyone have a suggestion on what I could try to fix this or how I could narrow down what the issue might be? I have a Dionex 5000+ RFIC system, that has a suppressor, a continuously regenerated anion trap, and an added carbonate reducer. I am also currently doing auto-injections.
Thanks for any suggestions.
Hi Kuju,

Welcome to the forum. Based on what you have described- negative dips immediately after all analyte peaks and overall reduced response- it sounds like your suppressor capacity may be slightly depleted. Please try the chemical regeneration described in section 5.2.1 (steps A-C on page 77) of the suppressor manual (link here).
Best regards,

John Guajardo
Senior Product Manager
Thermo Fisher Scientific
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry