Hello

I have question, can I change mobile phase flow rate eg. from 0,800 ml/min to 0,700 ml/min? will this change be affected by the calculation of the concentration? My calibration curve was provide on 0,800 ml/min.

Thanks
You can certainly the flow rate but you WILL have to rerun the calibration at the new flow. Remember that peak area is determined by response multiplied by residence time in the detector cell.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thank you,

OK, I prepare new calibration curve .

I wish you all the best in New Year
You can run a standard (made in your lab) or CRM with a flow of 0.7ml/min, and keep the calibration curve with a flow of 0.8 ml/min, and you can see how much it affects the concentration.
It would be best for each flow to have a calibration curve.
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 0 registered, 0 hidden and 2 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 2 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry