Good morning to everyone

I would like to ask some question about the calibration of my IC.
I use Professional IC Vario ONE with SeS/PP and automatic dilution.

Does the order of standards affect calibration? E.g Is diference when I start mesure my calibration standards from the lowest concentration to the highest or from the highest to lowest? Because, when I start calibration from highest to lowest my calibration curve look like better and deviation between two IC machines is lower. I provide measurements in ppm range.

Thanks
Dear Samic

Typically I do not expect a difference in calibration curve depending on up or down calibration.

When the difference between the two calibration curves are significant, the reason should be searched.

Please send me the two sets of calibration runs (*.idet) for further discussion. e-mail: markus.laeubli@metrohm.com
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry