Hi All ,

I just start working HPAEC PAD ... first runs ok , +/- 100 samples ... suddenly no retention , only flatline ... I did'n t change the conditions ... What could be the reason ?
You have an electrical bad connection or a leak or...
Hi Harry64,

What is the condition of the electrode that you're using? These days there are "cards" where the electrode is swapped into and out of the detector cell, in earlier versions of the IPAD the electrodes had to be cleaned much in the same way as a MS source. How were the responses for standard injections...did they decrease in value as the sequence(s) progressed?
MattM
Hi all again ,

Working with HPAEC PAD ... bad respons on peak .. 3 x lower than expect .
Changed , disposable gold electrode , eluents ( new NaOH , H2O ... ) new reference electrode calibrated ... also working with inguard H system ...
still bad respons afterwards ...

What could be the cause of this ? Carbonates-contamination , column , guard column ,

many thanks for answer ...
Hi Harry64,

Please contact me offline (john.guajardo@thermofisher.com) and I will help you with this issue. Please include a Chromeleon cmb(x) backup file of a few recent injections or sequence when everything was working properly vs. now with the decreased response.
Best regards,

John Guajardo
Senior Product Manager
Thermo Fisher Scientific
John Guajardo wrote:
Hi Harry64,

Please contact me offline (john.guajardo@thermofisher.com) and I will help you with this issue. Please include a Chromeleon cmb(x) backup file of a few recent injections or sequence when everything was working properly vs. now with the decreased response.


Hi John , I send you the chromeleon file tomorrow in the morning . many thx in advance ...rgrds H64
6 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry