Hi,

I have a problem with my Dionex DP (ICS-3000) pump. I get a warning message saying "Drive Current too large. Camshaft xx" with the xx being an angle. Then the pump stops automatically when I get the message "Excessive Drive current. Camshaft xx".

The manual says that this could be due to blocked tubes but I have changed all of the tubing and I still get the messages. I even changed the whole pump mechanism but still get the messages.

Can anyone please help me? I am very much in need of a working pump.

Thanks
Hi,

this is most probably a defective pump mechanism.
You will have to get a Thermo Fisher Scientific SVCRep onsite to exchange it, along with FW update to latest version.


Greetings
Had anything changed prior to receiving the error message? Changes in hardware, configuration, etc.?

How confident are you that you installed the new pump mechanism properly and that there isn't binding of the camshaft?
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry