Peak area problems

Posted: Thu Jan 26, 2017 11:42 pm
by gnikol21
Hello to all,
Im working in an analysis of a cation and I have problem with the repeatability of the peak area of my analyte between a series of injections.
Im using a system with an IC detector, chemical suppressor and an HPLC waters. I know that in IC needed a full plastic system. The HPLC part in my system contains stain steel parts which cannot be replaced with plastic ones.
Is there any possibility the stain steel parts be the cause of the peak area variability?

Thanx in advance

Re: Peak area problems

Posted: Fri Jan 27, 2017 12:21 am
by tom jupille
Is there any possibility the stain steel parts be the cause of the peak area variability
*Anything* is possible, but unless you tell us what the analyte/mobile phase/column are, there is no way to say any more.

Re: Peak area problems

Posted: Sun Jan 29, 2017 10:32 pm
by gnikol21
Well, the analyte is the choline cation, the mobile phase is methanesulphonic acid (6mM) and im using the column IonPac CS19 with a relevant guard.

Re: Peak area problems

Posted: Sun Jan 29, 2017 11:04 pm
by tom jupille
Those should be compatible with stainless steel.

Re: Peak area problems

Posted: Thu Feb 02, 2017 3:48 pm
by Markus Laeubli, Metrohm
I would recommend to have a metal free system when working with MSA as an eluent.

Re: Peak area problems

Posted: Thu Feb 02, 2017 8:09 pm
by DR
Could the problem be that you're using plastic syringes and filters for your final sample prep step and introducing semi-random amounts of chloride?

Is there a pattern to the variations in area? If areas are generally drifting up or down, you might have to consider chilling your samples as they may not be stable.