Differences IC25 vs DX-120
Posted: Sun Jul 10, 2016 9:06 pm
by lespaul
Hi everyone,
Does anyone know the biggest differences between the Dionex IC25 and the DX-120 chromatograph systems (IC).
I know their both integrated systems but the specs looks quite similar. The IC25 is a bit newer, but I cant see any major differences.
DX-120 manual -
http://www.dionex.com/en-us/webdocs/4520-31183-03.pdfIC25 manual -
http://www.dionex.com/en-us/webdocs/4536-31685-01.pdf
Re: Differences IC25 vs DX-120
Posted: Mon Jul 11, 2016 9:10 am
by sbrand
Hi Lespaul
There are several differencies in concept and technology:
The DX120 is an integrated system with pump, detector injector and column/suppressor space. But the pump is a single piston pump only. If you have software or a chart recorder you can start with IC.
The IC25 (also know as DX320) is a combined pump/detector modul. The pump is a dual piston pump (electronically controlled) and i would say more flexibel/precise. But there is no compartement for the column or suppressor and no inject valve. You must use a single injector (stand alone) or an autosampler.
Both system are isocratic and are out of production since several years (and service can only be done on a "best effort" base).
If you have to analyze only the so called standard anions or cations and no need for an autosampler you may better start with the DX-120.
Hope this helps.
Best regard
Re: Differences IC25 vs DX-120
Posted: Fri Jul 15, 2016 11:28 pm
by duck duck goose
Just one detail to clarify, you are going to need an autosampler for either system. Also, from what I have seen, there seem to be more IC25 units still in use than DX-120 units. If you are looking for longevity from either of these systems you might have an easier time sourcing parts on the IC25 rather than the DX-120.
Re: Differences IC25 vs DX-120
Posted: Mon Jul 18, 2016 9:17 am
by sbrand
@duck duck goose: The DX-120
includes a rheodyne inject valve with a luer port adapter at the front door. Therefore there is no need to use an autosampler. You can direct load the sample manually with a syringe. With the DX-320 you need either a manual inject valve or an autosampler. The DX-120 can be easily equipped with a an commercial available autosampler (e.g.
http://www.dionex.com/en-us/products/ion-chromatography/ic-rfic-modules/autosamplers-injectors/lp-72578.html.
Hope this helps.
Cheers