Peak phosphate getting small and broad (ICS-5000)

Discussions about IC and related topics

5 posts Page 1 of 1
Hello,

What to do when phosphate peak getting broad and small?
We used around 0.1-0.4ppm as standard 1,2 and 3 respectively.
The recovery for std 1 and 2 are not good. Thus affect the Coeff. Of determination of phosphate.
But there is no problem with micropipette.
We control our Coeff. Of Det. 99.5% above.
We also control the slope and fixed the offset to zero.
From our observation the Coeff.Det. and slope getting decreases and the peak look broad and small (standard 2 & 3). We want high and sharp peak.
We already used new suppressor.
But after use it for 3 weeks, the problem occurs for phosphate only.
Other ions are within control except phosphate.
What should we do?

Thanks.
When phosphate is getting too small usually a contamination of the IC with iron is the reason.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Thank you.
But how to solve this matter?
We already clean the suppressor with IPA, 0.5N H2S04 and purified water.
But the peak still broad and small.
Then we change the concentration of eluent generator.
The peak became sharp but relative area and height of the peak is lower than usual
caused the slope of phosphate is still in lower site of our control limit.
Anyone have any suggestion?

Thanks.
On Metrohm IC systems we use sulfuric acid/oxalic acid/ acetone regeneration of the Metrohm Suppressor Module. Check with your service rep about possibilities.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Thanks for sharing.
Actually our IC system is still ok because we still can get Coeff. Of Det. 99.5% above and ICV recovery within 80-120% but we have to control the offset and slope. When the offset or slope of any element is out of control limit, we have to take action in order to get all within control limit. Right now, we always having slope issue. We have to think the solution by our self and ask anyone instead of calling service engineer or buying spare part in order to reduce our budget.
Did anyone struggle like us?
How to control slope?
It is really difficult to control all things.
:cry:
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry