Baseline issues

Discussions about IC and related topics

2 posts Page 1 of 1
I am having a problem with our IC method, specifically nitrate. All other QC's pass but nitrate is consistently ~60% of what it is supposed to be. We are running a waters 2695 with Waters 432 suppressed conductivity detector, and a dionex ionpac as14 column. The last 6 peaks are the ones of interest, Standard 3 should be close to the same size as QC, and they are for all peaks except nitrate. I have tried different sets of standards, and different sets of QC's multiple times and I keep getting similar results. Any input?

Image
Hi Mritari,

What are you using as a suppressor--perhaps the SeQuant one? Could the suppressor itself be fouled with some component(s) that are present in the QC matrix but not the standard? Or from something from within the samples...such as iron or manganese?

Please, see what you think, and best wishes.
MattM
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry