Residual pulsation on IC pump [Metrohm]

Discussions about IC and related topics

2 posts Page 1 of 1
Hi Dear Friends!
We have 881 Compact IC pro Anion MCS, please answer question.

Question: If use manual, technical specifications=>High pressure pump=>Residual pulsa-
tion< 1 %, but we have 8.0 - 8.5 MPa fluctuations after completely equilibrated system (>60 min), temperature of column heater - 50 C.
sorry for the late answer.

A pulsation as you mention it is by far to high for a proper running system. You actually speak about fluctuations. Also for baseline stability the range is too high.
But if you speak about pulsation you need to look for regular pressure changes in the frequncy of the piston mouvment.

I strongly recommend to check back with the Metrohm Service.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry