I'm getting high variability in fluoride analysis (AS15 on ICS-5000). Other analytes work fine. Any thoughts?
-
- Posts: 310
- Joined: Mon Aug 30, 2004 4:56 pm
- Location: Sunnyvale California USA
iceyuser,
High variability of retention time or area?
Chris
High variability of retention time or area?
Chris
-
- Posts: 15
- Joined: Sat Mar 01, 2014 5:28 pm
Chris, it's variability in the peak area. The retention time is consistent.
-
- Posts: 310
- Joined: Mon Aug 30, 2004 4:56 pm
- Location: Sunnyvale California USA
iceyuser,
The most likely causes of this are setting the peak width too wide or having the detector data rate too slow. You should be at 5Hz and 0.1 minute minimum peak width.
Chris
The most likely causes of this are setting the peak width too wide or having the detector data rate too slow. You should be at 5Hz and 0.1 minute minimum peak width.
Chris
-
- Posts: 5
- Joined: Fri May 03, 2013 4:33 am
Hi iceyuser,
Check your Matrix as well, chelating metals such as Alumina or high Iron concentrations can influence fluoride recovery
Mark Albertson
Thermo Fisher Scientific
Sales Mgr ANZ
Check your Matrix as well, chelating metals such as Alumina or high Iron concentrations can influence fluoride recovery
Mark Albertson
Thermo Fisher Scientific
Sales Mgr ANZ
-
- Posts: 1
- Joined: Sun Oct 06, 2013 6:38 pm
On a similar topic, can anyone recommend the best column for improving fluoride resolution from the water dip? Currently using Dionex IC20 with #051786 AS9-HC high capacity column, with a 9mM carbonate eluent. Has anyone had experience with Allsep anion IC columns? Or is there another good way to improve fluoride without the extra step of spiking all the samples with carbonate to eliminate the dip?
-
- Posts: 310
- Joined: Mon Aug 30, 2004 4:56 pm
- Location: Sunnyvale California USA
You don't say whether you're open to switching to hydroxide mobile phases but, if so, a good choice would be the IonPac AS19. If you want to stick with carbonate eluent systems the IonPac AS23 is roughly equivalent to the AS9-HC but has superior resolution of fluoride from the void.
It's been a while since I evaluated an Allsep column but in my experience this column isn't state-of-the-art. While they still sell IC columns such as the Allsep column, Grace exited the IC instrument business years ago and hasn't done any IC column development work since.
Showdex has a number of competitive products with good fluoride including the IC SI-50 4E that you might consider as well.
It's been a while since I evaluated an Allsep column but in my experience this column isn't state-of-the-art. While they still sell IC columns such as the Allsep column, Grace exited the IC instrument business years ago and hasn't done any IC column development work since.
Showdex has a number of competitive products with good fluoride including the IC SI-50 4E that you might consider as well.
-
- Posts: 3475
- Joined: Mon Jan 07, 2013 8:54 pm
- Location: Western Kentucky
A Metrohm ASupp 7 column does a pretty good job of separating Fluoride from the water dip when using a carbonate eluent. I don't remember off hand the eluent concentration but they have several applications posted online there.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 362
- Joined: Tue Aug 31, 2004 6:54 am
Standardeluent for the Metrosep A Supp 7 - 250/4.0 is 3.6 mmol/L sodium carbonate.
See AN-S-256: Thirteen anions separated on the Metrosep A Supp 7 – 250
For faster separations Metrosep A Supp 7 - 150/4.0 is available.
See AN-S-256: Thirteen anions separated on the Metrosep A Supp 7 – 250
For faster separations Metrosep A Supp 7 - 150/4.0 is available.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
-
- Posts: 2
- Joined: Fri Sep 04, 2015 6:21 pm
Have had good luck running fluoride with a Metrohm A Supp 5-100 x4mm column with 3.2mM Na2CO3/1.0mM NaHCO3 eluent. Solved a similar reproducibility issue by cleaning the suppressor with 2,6 pyridinedicarboxylic acid to remove chelating metal contamination. Just a thought.
10 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science