Hi, I am new in chromatography and I need an advice. I am making method for detecting chloride, nitrate and sulfate in rain. System is Termoscientific ICS 5000+ capillary, column AS11HC, flow rate 0.015 ml/min, temperature 30 C, KOH eluent generator in isocratic mode at concentration 7.5 mM, instrument method time 17.5 minutes.
I done validation tests (accuracy, precision....) and everything seems ok, except there is constant decreasing in retention time, especially for sulfates, in period of 3 weeks from 12.5 minutes to 11.8 minutes and still decreasing. Retention time for chloride was 3.6, now is 3.5, nitrate was 7 minutes now ist 6.8.
Pressure/flow rate is stable, what can be a problem? Is it that maybe concentration of KOH eluent (7.5 mM) is too low for this method (column)?

best regards
Retention times always decrease as a column ages. I have had a couple of bad AS-18 columns that have had them shift pretty rapidly. I just had to recalibrate every other week and lower the eluent concentration to compensate. I have an Integrion with an AS-18 fast column and sulfate shifted earlier after a few months. The other anions are more stable. I compensated for that by increasing the column temperature. For some reason sulfate is more sensitive to temperature than the other anions on an AS-18 fast. I managed to get 18 months out of it by gradually moving the temperature from 30 - 35°.
Dear anionman,
thank you for the answer and advice, for start I will also try to change a temperature and see if there will be positive effect on sulfate retention..

best regards
3 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry