Sciex Tune Report

Posted: Tue Aug 27, 2019 6:44 pm
by alexcowart
When reviewing a tune report I know that the mass shift should be less than .1 Da and that the resolution/width should be .7 Da or less.

Is there a specification for the intensity per scan? For example when I run PPG how am I to know if 4.93E5 cps for a given mass is a strong signal or a weak signal?

Additional information that may be helpful is that I am using a Sciex 3200 Q Trap.

Re: Sciex Tune Report

Posted: Thu Sep 05, 2019 7:10 pm
by alexcowart
Somebody has to know something!

Re: Sciex Tune Report

Posted: Thu Sep 05, 2019 8:37 pm
by James_Ball
alexcowart wrote:
Somebody has to know something!


I don't have it handy at the moment, but I always compare to the data packet they left with from the initial install of the instrument. I will have to see if I can find it when I get back to the LCMS again.

Re: Sciex Tune Report

Posted: Tue Sep 10, 2019 2:46 pm
by alexcowart
I would appreciate that greatly!

Re: Sciex Tune Report

Posted: Wed Sep 11, 2019 6:17 am
by bunnahabhain
James_Ball wrote:
alexcowart wrote:
Somebody has to know something!


I don't have it handy at the moment, but I always compare to the data packet they left with from the initial install of the instrument. I will have to see if I can find it when I get back to the LCMS again.


Go to the "API Instrument" project and open the data folder named "year-month-installation" (or similar). In this folder, you will find the data that was acquired during installation, you can use that as a reference for your instrument's performance.